Knuckledragger Posted May 30, 2021 Share Posted May 30, 2021 So I bought my Apex setup complete with Trident in March during the sale. I got everything else set up accept for the Trident. I set it up today (yes, stupid to start on a holiday weekend). It ran through the initial setup and then it said it could not draw a sample. I went through and checked to make sure the sample line and waste were clear and not kinked and ran it again. It kept saying it was initializing but also no sample water. I pulled the reagent tray out to find a puddle of reagent underneath and reagent dripping down the dosing heads. I am so upset and frustrated! Has anyone else had this issue? I suspect I will have to wait until Tuesday and then have them replace the defective unit. Quote Link to comment Share on other sites More sharing options...
SuncrestReef Posted May 30, 2021 Share Posted May 30, 2021 Pull out the reagent tray and look inside the Trident at the center manifold where each of the tubes connect. Look for any tubes disconnected from the fittings. If anything is disconnected, you can remove the Trident's outer shell to gain access to the fittings and reattach the tube(s). Quote Link to comment Share on other sites More sharing options...
Knuckledragger Posted May 30, 2021 Author Share Posted May 30, 2021 It looks like everything is hooked up I think. Here is a pic. Quote Link to comment Share on other sites More sharing options...
SuncrestReef Posted May 30, 2021 Share Posted May 30, 2021 You'll likely need to contact Neptune Support. But if you want to further investigate on your own, wipe up the spilled reagents with a paper towel, then re-run the initialization process with the tray removed so you can watch inside to see where the leak is coming from. Quote Link to comment Share on other sites More sharing options...
Danik Posted May 30, 2021 Share Posted May 30, 2021 This looks like it could be broken Quote Link to comment Share on other sites More sharing options...
Knuckledragger Posted May 30, 2021 Author Share Posted May 30, 2021 36 minutes ago, Danik said: This looks like it could be broken I agree. It looks like the tube should be pushed on the black plastic nipple like the others. Quote Link to comment Share on other sites More sharing options...
Knuckledragger Posted June 2, 2021 Author Share Posted June 2, 2021 So it was a collapsed waste line. Apparently this not to uncommon out of the box as I found multiple others online with the same issue. Flushed the line while rubbing between my fingers along the length of it and it started flowing through. I took the cover off and cleaned everything up and reconnected the tube that had popped off with tweezers, which is apparently what they were going to have me do in the support call. After running the initial setup again it finally started working and appears to be settling in on readings. The calcium shows way high at 655 but I posted a new thread about that issue. Quote Link to comment Share on other sites More sharing options...
titus'reef Posted June 2, 2021 Share Posted June 2, 2021 On 5/29/2021 at 10:45 PM, Danik said: This looks like it could be broken Good eye Danik Quote Link to comment Share on other sites More sharing options...